Threaded resistance thermometer Miniature design Model TR10-D

Similar documents
Threaded thermocouple Miniature design Model TC10-D

Resistance thermometer With spring-loaded tip Model TR55

Threaded resistance thermometer With protection tube Model TR10-C

Bayonet resistance thermometer Model TR53

Resistance thermometer For additional thermowell Model TR10-B

Threaded resistance thermometer Model TR10-C, with protection tube

Gas-actuated thermometer Stainless steel version Model 73

Reed sensor For bypass level indicators Model BLR

Hand-held thermometer, industrial version Model CTH6300 Model CTH63I0, Ex version

Hand-held thermometer, precision version Model CTH6500 Model CTH65I0, Ex version

Differential pressure sensor For ventilation and air-conditioning Model A2G-50

Thermocouples Model TC200, for Additional Thermowell

Miniaturised pressure sensor Model M-10, standard version Model M-11, version with flush diaphragm

Temperature dry-well calibrator Models CTD9100-COOL, CTD , CTD , CTD

Pressure sensor For highest pressure applications to 15,000 bar Model HP-2

Temperature dry-well calibrator Models CTD , CTD

Electronic pressure switch with display Model PSD-30, standard version Model PSD-31, with flush diaphragm

Micro calibration bath Models CTB , CTB

Level sensor Stainless steel version Model RLT-1000, for industrial applications

Temperature dry-well calibrator Models CTD9100-COOL, CTD , CTD , CTD

Level sensor with reed measuring chain Model FLR, for the process industry

Bourdon tube pressure gauge with electrical output signal Stainless steel, safety version, NS 63 Model PGT23.063

Hand-held thermometer Model CTH7000

Hand-held thermometer Model CTH6200

Dual differential pressure sensor For ventilation and air-conditioning Model A2G-52

Terminal Boxes Series 8118

Float switch For industrial applications Model RLS-1000

Precision digital pressure gauge Model CPG1500

Hand-held thermometer Model CTH7000

Programming unit VIATOR HART modems Model PU-H

Electronic pressure switch with display Model PSD-4

Electronic pressure switch with display For sanitary applications Model PSA-31

LMP 308. Separable Stainless Steel Probe. Stainless Steel Sensor. accuracy according to IEC 60770: standard: 0.35 % FSO option: 0.25 % FSO / 0.

Portable pressure calibrator Model CPH7600

Precision thermometer Model CTR2000

Multi-functional precision thermometer CTR3000

Pressure transmitter for highest pressure applications Up to 217,500 psi (15,000 bar) Model HP-2

Precision digital pressure gauge Model CPG1500

Electronic pressure switch with display For sanitary applications Model PSA-31

Float Switches with Permanent Magnet For Vertical Installation Model RSM

Precision thermometer Model CTR5000

CalibratorUnit Models CPU6000-W, CPU6000-S, CPU6000-M

DMP 333 DMP 333. Industrial Pressure Transmitter For High Pressure. Stainless Steel Sensor. Pressure Transmitter. Industrial

ProcessCalibrator Model CPH6000

HMT360 Series Intrinsically Safe Humidity and Temperature Transmitters

Load Disconnect Switches and Motor Switches

Electronic pressure switch with display For sanitary applications Model PSA-31

LMP 308i. Separable Stainless Steel Probe Precision. Stainless Steel Sensor. accuracy according to IEC 60770: 0.1 % FSO

Precision digital gas density indicator Model GDI-100-D

Precision hand-held pressure indicator Model CPH6400

CombiTemp TFRN/TFRH. Operating instructions. Temperature sensor. Operating Instructions: EN.

Level measurement. Applications. Special features. Description

Magnetic field sensor Magnetic-inductive Proximity Sensor BIM-M12E-Y1X

LMP 308i. Separable Stainless Steel Probe Precision. Stainless Steel Sensor. accuracy according to IEC 60770: 0.1 % FSO.

ProcessCalibrator Model CPH6000

Inductive dual sensor For rotary actuators NI4-DSU35TC-2Y1X2/S933

Float Switches with Permanent Magnet For Horizontal Installation Model HIF

Portable low-pressure controller Model CPC2000

Hand-held thermometer Model CTH6200

Grounding Monitoring Device Series 8485

Grounding Monitoring Device Series 8485

Magnetic float switch For vertical installation Model FLS

Inductive sensor Dual sensors with extended temperature range NI4-DSU35TC-2Y1X2/S97

LMP 307. Stainless Steel Probe. Stainless Steel Sensor. accuracy according to IEC 60770: standard: 0.35 % FSO option: 0.25 % / 0.

2.3. Ex-Linear light fittings for fluorescent lamps. ellk W 58 W / ellm W 36 W (Zone 1, 2, 21, 22)

HMT360 Series Intrinsically Safe Humidity and Temperature Transmitters

Inductive sensor with analog output BI15-M30-LI-EXI

Inductive sensor NI3-EG08K-Y1-H1341

USB pressure transmitter model CPT2500 with model CPA2500 USB adapter and USBsoft2500 software

Inductive sensor For rotary actuators NI4-DSU35TC-2Y1X2/S933

Precision digital pressure gauge Model CPG1500

Precision digital pressure gauge Model CPG1500

Cable entry systems for potentially explosive atmospheres

Standard reference resistor Model CER6000

Temperature Measurement SITRANS TS300

2.2. Ex-Linear LED light fittings. ellk/m 92 LED 400/800 / ellk/m 92 LED 400/800 V-CG-S / NE / NIB-HT / NIB-CSA (Zone 1, 2, 21, 22)

THERMASGARD TM 65. Immersion screw-in duct temperature measuring transducer, calibratable, with multi-range switching and active output TECHNICAL DATA

SPRING LOADED RESISTANCE THERMOMETER CONFIGURATIONS

Series 3730 Type Electropneumatic Positioner

Inductive sensor BI1.5-EG08-Y1-H1341

Inductive sensor For rotary actuators NI4-DSU35TC-2Y1X2

USB pressure sensor With USB adapter and software Model CPT2500

Pneumatic Control/Feedback

T EN TROVIS Electropneumatic Positioner

Inductive sensor BI1.5-EG08-Y1-H1341

LS 6600 Technical Datasheet

Flow Measurement SITRANS F C

ExCos-P Pressure sensor 20 Pa Pa

Operating and monitoring systems Panel PC SERIES 400 Device platform EAGLE PM MT-436-TX

Operating and monitoring systems Panel PC SERIES 400 Device platform EAGLE PM ET-436-TX

Float switch For the process industry, lateral installation with external chamber Model ELS

21023, 21025, 2/2 Direct solenoid actuated poppet valves

Hand-held pressure indicator Model CPH6300-S1 (1-channel version) Model CPH6300-S2 (2-channel version)

Magnetic Field Sensor Magnetic-inductive Proximity Sensor BIM-M12E-Y1X-H1141

Approx. 4 s, measured in oil, with a probe Ø of 10 mm made of stainless steel. Influential effect of the ambient temperature

Pneumatic Control/Feedback

Digital Output Module 4-Channel Version for Zone 1 Series 9475/

SolConeX Wall-Mounted Socket, 16 A

Operating instructions (Safety-related part ATEX and IECEx) Inductive NAMUR sensors NN504A NN505A NN506A NN507A /00 08/2015

Inductive angle sensor with analog output Ri360P1-DSU35TC-ELi-Exi

Transcription:

Temperature Threaded resistance thermometer Miniature design Model TR10-D WIKA data sheet TE 60.04 for further approvals see page 2 Applications Machine building, plant and vessel construction Propulsion technology Air-conditioning and refrigeration systems Special features Sensor ranges from -196... +500 C (-320... +932 F) Compact design Universal application Direct installation into the process Explosion-protected versions Description Resistance thermometers of this series are used for the measurement of liquid and gaseous media at low and medium pressures. Fig. left: Model TR10-D with process connection compression fitting Fig. right: Model TR10-D with process connection double threaded hex bushing The resistance thermometer is screwed directly into the process. The electrical connection is made via connection terminals in the connection head (splash-proof). The measuring inserts are available in two variants, depending upon the application. There is a choice between versions with a replaceable, spring-loaded miniature measuring insert and versions with a non-replaceable measuring resistor built directly into the thermowell tip. Insertion length, process connection and sensor can each be selected for the respective application. Page 1 of 10 Data sheets showing similar products: Thermocouple for additional thermowell; model TC10-B; see data sheet TE 65.02 Cable resistance thermometer; model TR40; see data sheet TE 60.40

Explosion protection (option) The permissible power, P max, as well as the permissible ambient temperature for the respective category can be seen on the EC-type examination certificate and the certificate for hazardours areas or the operating instructions. Approvals (explosion protection, further approvals) Logo Description Country EU declaration of conformity EMC directive 1) EN 61326 emission (group 1, class B) and immunity (industrial application) RoHS directive ATEX directive (option) European Union - Ex i Zone 0 gas [II 1G Ex ia IIC T1... T6 Ga] Zone 1 gas [II 2G Ex ia IIC T1... T6 Gb] Zone 20 dust [II 1D Ex ia IIIC T125... T65 C Da] Zone 21 dust [II 2D Ex ia IIIC T125... T65 C Db] IECEx (option) International (in conjunction with ATEX) - Ex i Zone 0 gas [Ex ia IIC T1... T6 Ga] Zone 1 gas [Ex ia IIC T1... T6 Gb] Zone 20 dust [Ex ia IIIC T125... T65 C Da] Zone 21 dust [Ex ia IIIC T125... T65 C Db] EAC (option) - Ex i Zone 0 gas [0 Ex ia IIC T3/T4/T5/T6] Zone 1 gas [1 Ex ib IIC T3/T4/T5/T6] Zone 20 dust [DIP A20 Ta 65 C/Ta 95 C/Ta 125 C] Zone 21 dust [DIP A21 Ta 65 C/Ta 95 C/Ta 125 C] INMETRO (option) - Ex i Zone 0 gas [Ex ia IIC T3... T6 Ga] Zone 1 gas [Ex ib IIC T3... T6 Gb] Zone 20 dust [Ex ia IIIC T125... T65 C Da] Zone 21 dust [Ex ib IIIC T125... T65 C Db] NEPSI (option) - Ex i Zone 0 gas [Ex ia IIC T3 ~ T6] Zone 1 gas [Ex ib IIC T3 ~ T6] KCs - KOSHA (option) - Ex i Zone 0 gas [Ex ia IIC T4... T6] Zone 1 gas [Ex ib IIC T4... T6] - PESO (option) - Ex i Zone 0 gas [Ex ia IIC T1... T6 Ga] Zone 1 gas [Ex ib IIC T3... T6 Gb] DNOP - MakNII (option) - Ex i Zone 0 gas [II 1G Ex ia IIC T3, T4, T5, T6 Ga] Zone 1 gas [II 2G Ex ia IIC T3, T4, T5, T6 Gb] Zone 20 dust [II 1D Ex ia IIIC T65, T95, T125 C Da] Zone 21 dust [II 2D Ex ib IIIC T125... T65 C Db] 1) Only for built-in transmitter Eurasian Economic Community Brazil China South Korea India Ukraine Page 2 of 10

Logo Description Country GOST (option) Russia Metrology, measurement technology KazInMetr (option) Metrology, measurement technology Kazakhstan - MTSCHS (option) Permission for commissioning BelGIM (option) Metrology, measurement technology UkrSEPRO Metrology, measurement technology Uzstandard (option) Metrology, measurement technology Kazakhstan Belarus Ukraine Uzbekistan Instruments marked with ia may also be used in areas only requiring instruments marked with ib or ic. If an instrument with ia marking has been used in an area with requirements in accordance with ib or ic, it can no longer be operated in areas with requirements in accordance with ia afterwards. Approvals and certificates, see website Page 3 of 10

Sensor Measuring element Pt100, Pt1000 1) (measuring current: 0.1... 1.0 ma) 2) Connection method Single elements Dual elements 1 x 2-wire 1 x 3-wire 1 x 4-wire 2 x 2-wire Accuracy class / Range of use of the sensor per EN 60751 Class Sensor construction Wire-wound Thin-film Class B -196... +500 C -50... +500 C Class A 3) -100... +450 C -30... +300 C Class AA 3) -50... +250 C 0... 150 C 1) Pt1000 only available as a thin-film measuring resistor 2) For detailed specifications for Pt100 sensors, see Technical information IN 00.17 at www.wika.com. 3) Not with 2-wire connection method The table shows the temperature ranges listed in the respective standards, in which the tolerance values (class accuracies) are valid. For detailed specifications for Pt100 sensors, see Technical information IN 00.17 at www.wika.com. Electrical connection 1 x Pt100, 2-wire 1 x Pt100, 3-wire 1 x Pt100, 4-wire 2 x Pt100, 2-wire black yellow 3383942.03 white white white white white white white white white black yellow For the electrical connections of built-in temperature transmitters see the corresponding data sheets or operating instructions. Page 4 of 10

Connection head JS Model Material Cable entry thread size Ingress protection Cap Surface Connection to (max.) 1) neck tube JS Aluminium M16 x 1.5 2) IP65 Cover with 2 screws Blue, lacque 3) M24 x 1.5, ½ NPT Model Explosion protection Without Ex i (gas) Zone 0, 1, 2 JS x x x 1) The ingress protection refers to the connection head, for information on the cable glands, see page 7 2) Standard 3) RAL 5022 Ex i (dust) Zone 20, 21 Cable entry Standard Plastic Plastic (Ex) Brass, nickel-plated The pictures show examples of connection heads. Cable entry Cable entry thread size Standard cable entry M16 x 1.5 Plastic cable gland M16 x 1.5 Brass cable gland, nickel-plated M16 x 1.5 Cable entry Colour Ingress Min./max. ambient Explosion protection protection (max.) temperature without Ex i (gas), zone 0, 1, 2 Standard cable entry Blank IP65-40... +80 C x x Plastic cable gland Black or grey IP66, IP68-40... +80 C x - Plastic cable gland, Ex e Light blue IP66, IP68-20... +80 C (standard) x x -40... +70 C (option) Plastic cable gland, Ex e Black IP66, IP68-20... +80 C (standard) x - -40... +70 C (option) Brass cable gland, nickel-plated Blank IP66, IP68-40... +80 C x - Brass cable gland, nickel-plated, Ex e Blank IP66, IP68-40... +80 C x x Page 5 of 10

Ingress protection to IP65 per IEC/EN 60529 under the following conditions: Use of a suitable cable gland Use of a cable cross-section appropriate for the gland or select the appropriate cable gland for the available cable Adhere to the tightening torques for all threaded connections Transmitter (option) Within the model JS connection head, a model T91.20 analogue temperature transmitter can be factory-fitted. It is mounted in place of the terminal block. The version with temperature transmitter is not suitable for use in hazardous areas. For further specifications on the model T91.20 temperature transmitter please refer to WIKA data sheet TE 91.01. Transmitter model Output signal 4... 20 ma Transmitter (selectable versions) Model T91.20 Data sheet TE 91.01 Output 4... 20 ma x Connection method 1 x 2-wire x Measuring current 0.8... 1 ma 1) Explosion protection - Possible mounting positions for transmitters Connection head T91.20 JS Mounted instead of terminal block For a correct determination of the overall measuring deviation, the sensor and transmitter measuring deviations must be added. Page 6 of 10

Components model TR10-D Dimensions in mm Without process connection Double threaded hex bushing Extended process connection Threaded connection (welded) Compression fitting (adjustable) 11346418.02 3157966.01 Legend: Connection head Process connection Thermowell Terminal block Thread Legend: A (U 1 ) Insertion length (with parallel threads) A (U 2 ) Insertion length (with tape threads) N (M H ) Neck length (U) Insertion length per DIN 43772 (U 1 ) Insertion length in accordance with international standards Ø d Measuring insert diameter Page 7 of 10

Thermowell / Process connection Diameter 6 mm 8 mm Process connection Thread size Neck length (standard) without - 7 mm (Hex height) Double threaded hex bushing (male thread direct on the connection head) Male thread (offsetwelded to thermowell) Compression fitting with metal clamping ring Compression fitting with PTFE clamping ring 2) Spring-loaded compression fitting G ¼ B G ⅜ B G ½ B M10 x 1 1) M14 x 1.5 M18 x 1.5 M20 x 1.5 ¼ NPT ½ NPT Max. neck length Min. insertion length Max. insertion length N (M H ) N (M H ) A (U 1 ) / A (U 2 ) A (U 1 ) / A (U 2 ) 10 mm (Hex height incl. dimension up to the screw-in plane) approx. 19 mm (Hex height incl. dimension up to the screw-in plane) 7 mm (Hex height) 10 mm (Hex height incl. dimension up to the screw-in plane) approx. 19 mm (Hex height incl. dimension up to the screw-in plane) Material 50 mm 600 mm 1.4571 G ¼ B 55 mm 200 mm 50 mm 600 mm (incl. G ⅜ B G ½ B M10 x 1 1) M14 x 1.5 M18 x 1.5 M20 x 1.5 ¼ NPT ½ NPT neck length) G ¼ B G ⅜ B G ½ B M10 x 1 1) M14 x 1.5 M18 x 1.5 M20 x 1.5 ¼ NPT ½ NPT G ¼ B G ⅜ B G ½ B M14 x 1.5 M18 x 1.5 M20 x 1.5 ¼ NPT ½ NPT approx. 55 mm approx. 100 mm 1) only Ø = 6 mm 2) Maximum temperature at process connection: 150 C Page 8 of 10

Compression fitting Ferrules from stainless steel are only adjustable once; once the fitting has been loosened, sliding along the thermowell is no longer possible. Ferrules from PTFE can be adjusted numerous times; once the fitting has been loosened it can again be tightened onto the thermowell. Max. temperature at process connection: 150 C Operating conditions Ambient and storage temperature -40... +80 C Other ambient and storage temperatures on request On delivery, the compression fittings are only tightened hand-tight. Insertion length A and neck length N (M H ) can thus be checked. The final positioning/fixing of the compression fitting is carried out at the installation location. Neck length N (M H ) The neck length depends on the intended use. Usually an isolation is bridged by the neck tube. Also, in many cases, the neck tube serves as a cooling extension between the connection head and the medium, in order to protect a possible built-in transmitter from high medium temperatures. Measuring insert Specifications Description Diameter (for thermowell Ø = 6 mm) Diameter (for thermowell Ø = 8 mm) Operating temperatures (dependent upon the sensor design type and the accuracy class) Built-in measuring insert model Removable design The measuring insert is spring-mounted with two screws into the connection head and can simply be removed from the thermowell for calibration purposes. The thermowell itself can thus remain in the process. The terminal base for electrical connection is connected to the probe tube of the measuring insert. 3 mm - 6 mm - Min: -196 C Max: +500 C TR10-A (from 100 mm measuring insert lengths) Mineral-insulated line (MI cable) TR11-A (up to 99 mm measuring insert lengths) Tubular design Fixed design There is no removable measuring insert in this version. Instead, the sensor element is mounted directly in the thermowell tip. The terminal base for the electrical connection is permanently screwed into the connection head. Min: -50 C Max: +250 C - Page 9 of 10

Certificates (option) Certification type Measurement accuracy 2.2 test report x x 3.1 inspection certificate x x DKD/DAkkS calibration certificate x - Material certificate The different certifications can be combined with each other. Ordering information Model / Version / Measuring insert / Explosion protection / Process connection / Version and material of threaded connection / Thread size / Measuring element / Connection method / Temperature range / Design of the probe tip / Probe diameter / Insertion length A / Neck length N(M H ) / Certificates / Options 05/2008 WIKA Alexander Wiegand SE & Co. KG, all rights reserved. The specifications given in this document represent the state of engineering at the time of publishing. We reserve the right to make modifications to the specifications and materials. 07/2018 EN Page 10 of 10 WIKA Alexander Wiegand SE & Co. KG Alexander-Wiegand-Straße 30 63911 Klingenberg/Germany Tel. +49 9372 132-0 Fax +49 9372 132-406 info@wika.de www.wika.de